sddsexperiment varies process variables and measures process variables, with optional statistical analysis.
An input file of namelist commands gives the specific instructions.
The results are recorded in one or more SDDS files.
sddsexperiment LTP:H1.exp LTP:H1.sddswhere the contents of the file
LTP:H1.exp are
&measurement control_name = "LTP:PH1:x",
column_name="LTP:PH1:x", units=mm,
number_to_average = 10,
&end
&variable control_name = "LTP:H1:CurrentAO",
column_name="LTP:H1:CurrentAO"
! the corrector is varied in 5 steps from -1.0 to 1.0 amps.
index_number = 0, index_limit = 5,
initial_value = -1.0, final_value = 1.0,
&end
&execute
post_change_pause=4,
intermeasurement_pause=1
&end
where the line starting with a ``!'' is a comment.
usage: sddsexperiment <inputfile> [<outputfile-rootname>] [-echoinput] [-dryrun] [-summarize] [-verbose] [-ezcaTiming=<timeout>,<retries>]
The input file consists of namelist commands that set up and execute the experiment. The functions of the commands are described below.
variable -- specifies a process variable to vary, and the range and steps of the variation. More than one
variable command may be defined, so that many process variables may vary at a time.
When an arbitrary sequence of setpoint values is required, the setpoints can be read in
from a file.
measurement -- specifies a process variable to measure at each step during the experiment.
Instead of using many measurement namelist,
one can optionally use a sddsmonitor-compatible file.
execute -- start executing the experiment. One group of variable, measurement and execute commands
may follow another in the same file for multiple experiments.
erase -- deletes previous variable or measurement setups.
list_control_quantities -- makes a cross-reference file for process variable names and column names of the data file.
system_call -- specifies a system call (usually a script) to be executed either before a measurement or before
setting a process variable.
The following text describes all the namelist commands and their respective fields in more detail. The command definition listing is of the form
&<command-name>
<variable-type> <variable-name> = <default-value>
&end
where the part <variable-type>, which doesn't appear in an actual command, is used to illustrate
the valid type of the value. The three valid types are:
double -- for a double-precision type, used for most physical quantity variables,
long -- for an integer type, used for indexes and flags.
STRING -- for a character string enclosed in double quotes.
&<command-name>
[<variable-name> = <value>,]
&end
In the namelist definition listings the square brackets denotes an optional component. Not all variables need to be defined - the defaults may be sufficient. Those that do need to be defined are noted in the detailed explanations. The only variables that don't have default values in general are string variables.
variable
readback_attempts and readback_tolerance are defined with non-zero positive values.
When an arbitrary sequence of setpoint values is required (say a binary sequence),
the values can be read in from an SDDS file specified by the values_file field.
The fields of &variable associated for the range and steps are ignored in this case.
With multiple variable commands,
variables may be varied in a multi-dimensional grid. For example,
variables may be varied independently of each other, or some groups of
variables may vary together forming one axis of a multi-dimensional grid
(see item index_number).
&variable
STRING control_name = NULL
STRING column_name = NULL
STRING symbol = NULL
STRING units = "unknown"
double initial_value = 0
double final_value = 0
long relative_to_original = 0
long index_number = 0
long index_limit = 0
STRING function = NULL
STRING values_file = NULL
STRING values_file_column = NULL
long substeps = 1
double substep_pause = 0
double range_multiplier = 1
STRING readback_name = NULL
double readback_pause = 0.1
double readback_tolerance = 0
long readback_attempts = 10
long reset_to_original = 1
&end
control_name -- Required. Process variable name to vary.
column_name -- Required. Column name for the variable data recorded in the output file.
symbol -- Optional. Symbol field for the above column definition of the variable data.
units -- Optional. Units field for the above column definition of the variable data.
initial_value -- Required. The initial value of the process variable in the variation.
final_value -- Required. The final value of the process variable in the variation.
index_number -- Required. The counter (or index) number with which
the defined variation is associated. In a sddsexperiment run, counters must
be defined in an increasing order with no gaps starting from counter 0. That is,
the first variable command of the
file must have index_number = 0. The second variable command
must have index_number = 0 or 1. In the former case, the two variables
will move together with the same number of steps according their respective
initial_value and final_value. In the latter
case, the two variables will vary independently of each other with possibly different
number of steps in a 2-dimensional grid.
Counter number
is nested within counter
. Therefore it might be efficient
to assign devices with slower response times to
higher index_number counter.
index_limit -- Normally required.
Number of steps in the variation. Measurements are taken at each step.
When more than one variable is associated with the same counter, only the index_limit
of the first variable definition for that counter need to be defined.
If index_limit is defined in variable commands
of the same index_number value, then the first index_limit
remain in force.
relative_to_original -- Optional. If non-zero, then the variation range is defined
relative to the original process variable value (i.e. the value prior to running the program).
range_multiplier -- Optional. Factor by which the range, final_value - initial_value, is multiplied.
New values of initial_value and final_value are calculated while keeping the midpoint of
the range the same.
function -- Optional. A string of rpn operations used to transform the range specified
by initial_value, final_value, and index_limit.
For convenience, a few values are pushed onto the stack and are available
to the user: the original value of the process variable, and the
calculated grid value for the process variable on the current step or substep.
The calculated values are
recorded in the output file. The environment variable RPN_DEFNS
is used to read a rpn definition file at the start of the execution of sddsexperiment.
values_file -- Optional. An SDDS data file containing setpoints for the variable.
This is useful is one has arbitrary setpoints values to apply.
The values of the fields initial_value, final_value, _substeps,
range_multiplier and index_limit are ignored.
One can have other variable namelists with the same index_number
that don't use a file for the values.
The default index_limit of the other variable will be
set to the number of setpoint in the values file.
Thus the values in the file and the values calculated for the other variable
will vary together with the same number of steps.
values_file_column -- Required when values_file is specified.
values_file_column gives the column name of the setpoints data
in file values_file.
substeps -- Optional. If greater than one, the steps are subdivided into this number.
Measurements are not made at substeps. Substeps are useful
when the physical device associated with the process variable cannot reliably make
steps as large as those that might be defined with initial_value,
substep_pause -- Optional. Number of seconds to pause after the variable change of each substeps.
readback_name -- Optional. Readback process variable name
associated with control_name.
The default value for readback_name is control_name.
readback_tolerance -- Optional. Maximum acceptable
absolute value of the difference between the process variable
setpoint and its readback. A positive value is required in order to enable readbacks.
readback_pause -- Optional. Number of seconds to pause after
each reading of the readback_name process variable.
This pause time is in addition to other pauses defined.
readback_attempts -- Optional. Number of allowed readings
and readback pauses after a variable change has occured.
After this number of readings, the program exits.
The first readback is attempted immediately (i.e. no pause) after
sending a setpoint command to the control_name.
reset_to_original -- Optional. A value of 1 means
that the variable is reset to its original value when the
experiment terminates normally or abnormally.
measurement_file
&measurement_file
STRING filename = NULL
long number_to_average = 1
long include_standard_deviation = 0
long include_sigma = 0
double lower_limit = 0
double upper_limit = 0
long limits_all = 0
&end
filename -- Required. SDDS file containing measurement process variables.
The required and optional column definitions closely resembles those used in sddsmonitor.
The columns are:
ControlName -- Required string data for the process variable names.
ReadbackName -- Optional string data for the measurement data
columns names of the experiment output file. If the column is not present, then the
experiment output file's columns names uses the measurement process variable names themselves.
NumberToAverage -- Optional long data giving the number of measurements
to average at each step of the experiment. The average value of the measurements for each
process variable is written to the output file. Therefore individual readings are not recorded.
IncludeStDev -- Optional long data. If non-zero, then the standard
deviations of measurements are calculated and a column in the output file for this quantity
is generated.
IncludeSigma -- Optional long data. If non-zero, then the sigma
is generated.
LowerLimit, UpperLimit -- Optional double data. Must have
both or neither. Specifies a range outside of which the measurement
data is to be rejected, and the measurement be retaken.
If the number of invalid measurements (reset to 0 at each
measurement step) equals or exceeds the value
of allowed_limit_errors (default of 1) in command execute,
then the program aborts.
The average values written to the output file excludes measurements outside this range.
LimitsAll -- Optional long data. If non-zero for a particular PV, then
out-of-range data for this PV causes all other PVs to be remeasured.
By default, only the out-of-range PV is remeasured.
filename is read in. For those
optional columns above that are not defined, then the following measurement_file fields
will act as defaults (note the different spellings):number_to_average,
include_standard_deviation, include_sigma,
lower_limit, upper_limit, limits_all.
measurement
measurement_file, as both commands
merely adds to an internal list of measurement PV.
&measurement
long limits_all = 0
&end
control_name -- Required. Process variable name to measure.
column_name -- Required. Column name for the measurement data recorded in the output file.
symbol -- Optional. Symbol field for the above column definition of the measurement data.
units -- Optional. Units field for the above column definition of the measurement data.
number_to_average -- Optional. Number of readings to average. The output data
is the average value. The individual readings are not recorded.
include_standard_deviation -- Optional. Calculate the standard deviation
(a measure of the distribution) of the
measurement, and generate a column in the output file for this quantity.
include_sigma -- Optional. Calculate the sigma (error of the mean) of the
lower_limit, upper_limit -- Optional. Defines a range of validity for the
individual measurements. If the number of invalid measurements (reset to 0 at each
of allowed_limit_errors (default of 1) in command execute, then the program aborts.
limits_all -- Optional. If non-zero, then out-of-limits data for this PV causes
all PVS to be remeasured. By default, only the out-of-limit PV is remeasured.
execute
&execute
STRING outputfile = NULL
double post_change_pause = 0
double intermeasurement_pause = 0
double rollover_pause = 0
long post_change_key_wait = 0
long allowed_timeout_errors = 1
long allowed_limit_errors = 1
double outlimit_pause = 0.1
long repeat_reading = 1
double post_reading_pause = 0.1
double ramp_pause = 0.25
long ramp_steps = 10
&end
outputfile -- Optional. SDDS output file of variable and measurement data. The string
may contain the string ``%s'' which is substituted with the rootname value
of the command line. If rootname is not given, then the root of the intput file
is substituted . If outputfile is null, then the root of the input file
is used for the output file. See description of output file below.
post_change_pause -- Optional. Number of seconds to pause after each change before
making measurements.
intermeasurement_pause -- Optional. Number of seconds to pause between each measurement.
Individual measurements for averaging are taken at this interval.
rollover_pause -- Optional. Number of seconds to pause after a counter has reached
its upper limit, and must rollover to zero. This allows any physical devices
associated with the counter to settle after a change equal to the total range
of the variation.
post_change_key_wait -- Optional. If non-zero, then wait for a key press after
making variable changes but before taking measurements. A prompt is given.
allowed_timeout_errors -- Optional. Number of timeout ezca errors allowed before aborting the
program.
allowed_limit_errors -- Optional. Number of invalid range measurement errors
program. The valid range of a measurement is specified in the measurement command.
outlimit_pause -- Optional. Number of seconds to pause after an invalid range measurement error
occured. This is to permit equipment time to recover from whatever glitch caused the out-of-limit
reading.
repeat_reading -- Optional. The measurements and statistical analyses are repeated this number of
times for each variable settings. A row of data is written to the output file
for each repitition.
post_reading_pause -- Optional. Number of seconds to pause after taking a
set of measurements and making a statistical analysis.
If measurements are repeated then the pause is repeated after each set of measurements.
ramp_steps -- Optional. Number of steps in the variables PV ramp
which occurs at the start and the end of the experiment.
Ramping is necessary for some devices that do not respond well to large changes to their setpoints. Ramping is done at the start of the experiments to slowly change the variable PVs from their current values to their initial values. Another ramp is done at the end to slowly bring the variable PVs from their final values back the original values. Ramping back to original values is also done when the experiment aborts for some reason.
ramp_pause -- Optional. Time interval at each step of the variables PV ramp
which occurs at the start and the end of the experiment. This is not the same variable
as the pause between variable changes during the experiment.
erase
&erase
long variable_definitions = 1
long measurement_definitions = 1
&end
variable_definitions -- Optional. If non-null, then all the variable definitions are erased.
measurement_definitions -- Optional. If non-null, then all the measurement definitions are erased.
&list_control_quantities
&end
filename -- Required. Name of file. Columns defined are ControlName,
SymbolicName, and ControlUnits.
system_call
&system_call
STRING command = NULL
double post_command_pause = 0
double pre_command_pause = 0
long append_counter = 0
STRING counter_format = "%ld"
long call_before_setting = 0
long call_before_measuring = 1
STRING counter_column_name = NULL
&end
command -- Required. Name of shell command or script to execute.
index_number -- Required. Counter number with which the command will be associated. The command is executed
when this counter is advanced or rolled over.
index_number -- Optional. Number of times the command is executed for
the associated counter. This field is used only when the value of index_number above defines a new counter.
post_command_pause -- Optional. Number of seconds to pause after the completion of the command.
pre_command_pause -- Optional. Number of seconds to pause before executing the command.
append_counter -- Optional. If non-zero, the counter value is appended to the command when the
system call is made.
counter_format -- Optional. Format for the counter if the counter value is appended to the command.
call_before_setting, call_before_measuring -- Optional.
At a counter advance or rollover the command can be executed in one of three ways:
call_before_setting=1, call_before_measuring=1
call_before_setting=0, call_before_measuring=1
call_before_setting=0, call_before_measuring=0
counter_column_name -- Optional. If non-null, a column in the output file with this name is defined.
The values written to this column are the number of times the command had been called minus one. This
value doesn't rollover with its associated counter.
The output file contains one data column for each measurement and control process variable.
The names of these data columns are taken from the column_name field of the
respective measurement and variable commands. The data are written
as double precision floating point numbers.
In addition, some time columns and parameters are defined:
sddsplot to make the best coice of time unit conversions
for time axis labeling.
This file contains process variable names, readback column names, and units as string data. This data can be used for cross-referencing. The columns defined in this file are:
control_name field given in the
column_name field given in the
units field given in the
-echoinput -- echos input file to stdout.
-dryrun -- the ``variable'' process variables are left untouched during the execution. The ``measurement''
process variables are still read. The pauses are still in effect. System calls are not made.
-summarize -- gives a summary of the experiment before executing it.
-verbose[=very] -- prints out information during the execution such as notification of
setting and reading process variables. The option very prints out the average measurement values.
-ezcaTiming=<timeout>,<retries> -- sets EZCA timeout and retry parameters
-describeinput -- Printouts the list of namelist commands and fields of the input file.